A recall notice for Trilogy Evo O2 and Trilogy Evo Portable Ventilators supplied by Philips Electronics Australia Ltd as two software issues have been identified that relate to a pressure increase in the ventilators.
-
Infant/Pediatric External Flow Sensor (EFS) Calibration Pressure increase. An increase in the expiratory pressure (EPAP/PEEP) can occur when the paediatric/infant External Flow Sensor is used with an Active Flow or Dual Limb circuit and a manual circuit calibration is performed.
-
Pressure Drift (Continuous Usage). When a Trilogy Evo or Trilogy Evo O2 model is used continuously without any interruption to therapy, the baseline pressure may increase or decrease.
Philips Electronics Australia Ltd - Trilogy Evo O2 and Trilogy Evo Portable Ventilators
PRA number
2021/19171
Published date
8 Oct 2021
Product description
Trilogy Evo O2 and Trilogy Evo
Product codes: IN2100X15B and IN2110X15B
Multiple serial numbers (see attached serial number list)
ARTG 332200
(Philips Electronics Australia Ltd - Portable ventilator, electric)
Identifying features
Serial number list.pdf 296.49 KB
What are the defects?
Two software issues have been identified that relate to a pressure increase in the ventilators.
-
Infant/Pediatric External Flow Sensor (EFS) Calibration Pressure increase. An increase in the expiratory pressure (EPAP/PEEP) can occur when the paediatric/infant External Flow Sensor is used with an Active Flow or Dual Limb circuit and a manual circuit calibration is performed.
-
Pressure Drift (Continuous Usage). When a Trilogy Evo or Trilogy Evo O2 model is used continuously without any interruption to therapy, the baseline pressure may increase or decrease.
What are the hazards?
- Infant/Pediatric EFS Calibration Pressure Increase
Potential for inappropriate therapy to be provided when the EFS is used with manual calibration, increasing the risk of serious health complications including:
- Barotrauma (an injury to the lungs because of too much pressure)
- Hypotension (a decrease in blood pressure)
- Hypercarbia (too much carbon dioxide in the blood)
- Issue 2 – Pressure Drift (Continuous Usage)
Potential for inappropriate therapy to be provided to a patient if there is a pressure increase or decrease from the baseline pressure settings without alarm, increasing the risk of serious health complications including:
- Barotrauma (an injury to the lungs because of too much pressure)
- Hypotension (a decrease in blood pressure)
- Dyspnea (a sensation of shortness of breath)
- Hypoxemia (a low amount of oxygen in the blood)
What should consumers do?
Philips will be releasing a software correction for this issue. Philips will be contacting customers when the software is released. In the interim, customers are advised to refer to the steps provided in the Customer Letter (supplied to impacted users by Philips) until the updated software is installed on the device.
Infant/Pediatric EFS Calibration Pressure Increase
If you are using the infant/pediatric EFS, do not perform a manual circuit calibration. Instead, use the default calibration. The pressure increase will not occur if the default circuit calibration is used.
Pressure Drift (Continuous Usage)
If a pressure drift is detected, the decision to reset the device and return to its original, intended settings MUST be made in consultation with a clinician. Take steps to support the patient during a brief interruption in therapy, as needed.
For further information, consumers should contact Philips Customer Care Centre by phone on
1800 251 400.
Supplier
Philips Electronics Australia Ltd
Traders who sold this product
Hospital and health services in ACT, NSW, QLD and VIC
Where the product was sold
International
New South Wales
Queensland
Victoria
Australian Capital Territory
Responsible regulator