A recall notice for MyWay Harnesse supplied by Sunrise Medical Pty Ltd as the MyWay harness consists of two Velcro locking mechanisms. The first Velcro strap has reduced adhesiveness on the harness.
Sunrise Medical Pty Ltd - MyWay Harnesses
PRA No.
2020/18269
Date published
20 Mar 2020
Product description
MyWay Harnesses
Part: 173-4100-04, 173-4200-07, 173-4300-02, 173-4400-09,
173-4500-06, 173-4600-08
All lots and sizes
ARTG 298435
What are the defects?
The MyWay harness consists of two Velcro locking mechanisms. The first Velcro strap has reduced adhesiveness on the harness.
What are the hazards?
If the Velcro strap has reduced adhesiveness, it will not provide core stability or hold the user upright as intended, increasing the risk of a fall.
What should consumers do?
Sunrise Medical is advising consumers they may continue to use the existing harness until supplied with the new design, but pay attention to the status of the Velcro. Conduct daily visual checks of the harness, in particular, check the harness and all straps/buckles for signs of wear and tear.
Once consumers receive the replacement harness, it is essential that they return the old harness to their distributor in the supplied return addressed envelope.
For further information, consumers can contact Sunrise Medical Australia Customer Service
on 02 9678 6600.
Supplier
Where the product was sold
New South Wales
Northern Territory
Queensland
Victoria
Western Australia
Responsible regulator
Therapeutic Goods Administration is the responsible regulator for this recall.